NIMcalibration for Minitouner-S

PostPosted: Mon 25 Jan 2021 07:52
by EA2ARU
Good morning:
What values do you put in NIMcalibration for the Minitouner-S?

73,s de ea2aru.

Re: NIMcalibration for Minitouner-S

PostPosted: Mon 25 Jan 2021 08:14
by F6DZP
Hello,

Each NIM has its own calibration.
You must find it yourself.
- by receiving a DATV station on the higher direct frequency you can and setting the ppm correction to lock on the exact frequency
or best method :
- use the TSA(Tiny Spectrum Analyzer) software with a PPM detection procedure . With this method the ppm calibration can be written in the MiniTiouner Eeprom

Jean-Pierre F6DZP

Re: NIMcalibration for Minitouner-S

PostPosted: Mon 25 Jan 2021 08:37
by EA2ARU
Hello Jean-Pierre:
if it is possible, please, send the TSA(Tiny Spectrum Analyzer) . Or one direction for download.
73,s de Jabi, ea2aru.

Re: NIMcalibration for Minitouner-S

PostPosted: Mon 25 Jan 2021 12:46
by F6DZP
in the download section of this Forum.

If you download the Minitioune v0.9.9.1 package you will have inside the TSA software.

Run TSA software
Send a calibrated signal, a carrier to you MiniTiouner ( higher freq is better)
give the freq of the carrier in the Freq value of the SCANNER
press the PPM button
you must see the carrier in the back rectangle of graphic scan.
click getppm as many time you need to get the Led Green
when the led is green, click on write ppm button

that's all.

Jean-Pierre F6DZP